Egad! Thanks for catching that error Mike!
I have a copy of that procedure from an E-mail Wayne posted to the reflector
a year ago and it had the correct instruction to run CAL PLL on 40 meters
only! According to the e-mail he had just developed this procedure at that
time and the procedure was correct for the current firmware, so I never
thought to look at the one on the web site more closely.
Good catch!!
If anyone is confused about the CAL PLL issue, here's the original procedure
as published by Wayne August 20, 2003. It is correct:
-------------------------------------------------------------
This method relies on the following simple observation. If you tune in an
on-air
signal at a known frequency, the difference between the *measured* VCO and
BFO
(using CAL FCTR) *must* equal that frequency, or C22 is not set correctly.
(Actually, this holds for 160-17 m; on 15-10 m it's the sum, or VCO + BFO,
that
must equal the signal's frequency. But it's easier to do the adjustment of
C22
on 17 m or lower because you don't have to do any math at all, as I'll
explain
below.)
Here's the procedure. It requires revision 2.XX firmware, and assumes you
have
already done Alignment and Test, Part II, at some point. The K2 should also
be
allowed to come up to room temperature.
1. Tune in a signal at a known frequency. Use one that's at an *exact* kHz
boundary, so you can easily see when the VCO and BFO readings match in step
2.
(I use WWV at 10, 15, or 20 MHz.) Use USB or LSB mode rather than CW, so
that
there will be no CW receive offset. In the case of a K2 I was calibrating,
the
VFO read 10000.17 when the signal was tuned in perfectly. If it had read
10000.00, no further improvement would have been possible.
TIP: Zero-beat the carrier precisely, or listen to a voice signal and adjust
the
VFO for the best quality. The more accurately you tune in the signal, the
more
accurately you'll be able to set C22, below.
2. Run CAL FCTR. Now alternately move the K2's internal counter probe
between
TP1 (VCO) and TP2 (BFO), adjusting C22 in small increments until the kHz and
Hz
digits at the two test points match as closely as possible. In my case, the
two
readings matched at 14913.60 and 4913.60. The difference is exactly
10000.00--the frequency of the on-air signal.
3. Put the counter probe on TP1 (VCO), switch to 40 meters, and run CAL PLL.
4. Put the probe on TP2 (BFO) and run CAL FIL. For each operating mode, vary
each filter (or BFO) setting up 1 count, then back down, to force the K2 to
take
a new BFO measurement for each and store it in EEPROM.
The VFO dial should now be very well calibrated.
If we get a lot of positive feedback on this method, we'll post it as an
application note.
73,
Wayne
N6KR
--------------------------------------------
It's a mystery how the old CAL PLL procedure got put in there...
Ron AC7AC
_______________________________________________
Elecraft mailing list
Post to:
[hidden email]
You must be a subscriber to post to the list.
Subscriber Info (Addr. Change, sub, unsub etc.):
http://mailman.qth.net/mailman/listinfo/elecraft
Help:
http://mailman.qth.net/subscribers.htmElecraft web page:
http://www.elecraft.com