On Jun 7, 2007, at 9:50 AM, wayne burdick wrote:

> Brian Lloyd wrote:
>
>> ....being able to change the S-meter slope and intercept strikes  
>> me as being a bad option. To me that is like changing the  
>> calibration of a voltmeter or wattmeter because you like the  
>> needle pointer to move differently.
>
> I'd agree with you if there were a single world-wide standard for  
> S-9, and no need to compensate for slight differences in receive  
> gain from one unit to the next. But reality is that S-meters  
> usually require both scale and offset calibration.

Yes, they do. But they can be calibrated to a standard regardless of  
the gain of the radio.

> This is also more flexible. As I mentioned earlier, I set my S-
> meters up for 4 dB per S-unit. Here's why: I like a greater degree  
> of sensitivity in the S-meter so I can see the effects of things  
> like preamp on/off, filter changes, notch, NR, etc. It also makes  
> band-pass filters easier to tweak when there isn't a scope or AF  
> voltmeter handy, and you can more readily see the effect of an  
> improved antenna during A/B testing.

What you are doing is changing the calibration instead of changing  
the *resolution*. What you really want to do is to be able to resolve  
smaller changes easily. So blow up the scale. Add calibration points  
for half S-units. That would give you 3dB points on the meter. That  
is even better than your 4dB resolution!

> If hams wanted to be precise in assessing signal levels, we'd  
> report them in dBm and do a lot of averaging.

I agree. I would prefer to have a meter calibrated in dBm but we have  
used S-units for so long that it is part of the fabric. Heck, we  
still use the English Standard system of measurements in the US. I  
have to switch back and forth between metric and ES all the time. And  
sometimes it is convenient to measure resistance in ohms or  
conductance in mohs even though we know they are really the same thing.

> But for most operators this is a hobby, not a job  :)

It is a technical hobby. We measure voltage, resistance, current, and  
power to very accurate levels. Why should we therefore say that  
accurately measuring receive signal level is unimportant? You  
yourself say that you use the S-meter to:

"...see the effects of things like preamp on/off, filter changes,  
notch, NR, etc. It also makes band-pass filters easier to tweak when  
there isn't a scope or AF voltmeter handy, and you can more readily  
see the effect of an improved antenna during A/B testing."

Clearly you are using it as an instrument of measurement. Why not  
have it conform to a standard so that the readings are useful rather  
than just randomly relative?

I think this gets back to my comment about resolution. If you are  
using a quantized bar-graph display it is easier to change the  
calibration than to change the resolution. OTOH, three digits would  
be nice or even an analog meter. (I actually still prefer analog  
meters for a lot of things, especially doing calibrations involving  
tweaking things.)

Never mind.

73 de Brian, WB6RQN
Brian Lloyd - brian HYPHEN wb6rqn AT lloyd DOT com


_______________________________________________
Elecraft mailing list
Post to: [hidden email]
You must be a subscriber to post to the list.
Subscriber Info (Addr. Change, sub, unsub etc.):
 http://mailman.qth.net/mailman/listinfo/elecraft   

Help: http://mailman.qth.net/subscribers.htm
Elecraft web page: http://www.elecraft.com